Re: Funky Swissflow readings?
Did you set the T-Bal correctly? The divider is set to a fan's pulse rate output not the swissflows. Not sure how the T-Bal is setup but for the CF633 you just change the PPR setting to 1 instead of the standard 2 for a fan. Then whatever reading it puts out divide by 6100. I checked mine with a separate flow meter to verify.
Still though setting the divder to 1 would only give you 4000... Maybe the T-Bal is setup a different way.
Also did you put a filter in front of that Swissflow? Any little particle will ruin it. Swissflow recommends a 20micron filter.
|