Ok this is actually a work-related question but here goes:
I need to measure and log different parameters (temperature, pH, Eh, {Ca2+}, etc in a bioreactor we are building. It looks to me like the easiest way to do this is to just buy a NI DAQ card and use their free software to continuously log the data. But I am not turning up much information on using DAQ cards to capture readings from electrodes. Is this going to work ok? They output a mV reading over a pretty big range (pH 4 to 10 is a 350 mV range) so I would think that this would work fine. But is there a lot of additional smoothing/conditioning circuitry on a pH meter that I don't know about? Anyone have any reasons why trying this is a bad idea? The electrodes are all BNC so I am assuming it's relatively easy to mate them to the DAQ card's analog inputs...
__________________
Getting paid like a biker with the best crank... -MF DOOM |
![]() |
![]() |
![]() |
#2 |
Cooling Savant
Join Date: Nov 2002
Location: Portugal, Europe
Posts: 870
|
![]()
beware of noise on those low amplitudes.
Depends on the noise you get on the lines. Depends on the scale of the DAQ -> trying to read mV on a 0 to 5V analog DAQ, no good -> instrumentation amplifiers from mV to V. decent noise reduction, lots of products floating around (google ![]() several IC's from Maxim , Analog. http://www.analog.com/en/subCat/0,28...0%255F,00.html