'calibrating' the DUT instrument connections
as has been previously discussed by many,
the measurement connections to the DUT have their own 'losses' and these need to be quantified in order that the gross readings may be corrected to arrive at a net value for the particular DUT parameter some preliminary data was described in the thread on "C", this has been redone and expanded for these tests the inst crosses, RTD ends, and connecting tubing were insulated with 3/8" closed cell neoprene the hydraulic data is straightforward and the effect of temperature can be seen, though not really in the flow range of significance - below 1.5 gpm - with the instrumentation used of greater interest to me was the temperature data taken at the same time: the ambient temp was held at 25°C and the dT measured at 11.5, 5.5, and 0°C temperature difference; then run again at 15.3°C with 0°C difference there is a temp change w/o any DUT, for higher accuracy this must be addressed the WCing testing community might consider the use of these (cheap) crosses as a de-facto std, everyone would have the same corrections (there is an inst offset included in these curves which could be backed out) |
11-22-2005, 02:33 PM | #3 |
CoolingWorks Tech Guy Formerly "Unregistered"
Join Date: Dec 2000
Location: Posts: 2,371.493,106
Posts: 4,440
|
use a tee, easier
ftgs w/0-ring from McMaster-Carr Ben, if anything different is done to the crosses new cal curves are required if you, or anyone else, makes the crosses as described then these curves could be used not many will have the capability of generating these curves, why they are posted |
11-22-2005, 03:18 PM | #4 |
Responsible for 2%
of all the posts here. Join Date: May 2002
Location: Texas, U.S.A.
Posts: 8,302
|
Yep, I realize that (re: cal curves).
So... (schedule 40) 3/4 PVC tee, then 3/4" to 1/2" adapters (glued), then 3/8" ID tubing and 6" Silicone tubing. I like that setup. We're still going to have some variations, unless you can detail the setup here (i.e. protrusion of the pressure port inside the tee, protrusion of the temp probe and at what depth). Am I just being picky? When you state 6" OAL connection tubing, what precisely does that mean? Does that include the barb overlap? Bill, you picked this arrangement, but I can live with a schedule 40 tee, with a tapped pressure port as you have (I'm assuming it doesn't protrude inside the tee?). I would have proposed schedule 80 parts, which are threaded. Should we go into ID variations of barbs, or agree on a single spec and/or supplier? Or ignore as irrelevant? |
11-22-2005, 04:50 PM | #5 |
CoolingWorks Tech Guy Formerly "Unregistered"
Join Date: Dec 2000
Location: Posts: 2,371.493,106
Posts: 4,440
|
I'll gin up a BOM tomorrow, off the shelf stuff/McMasters-Carr
the previous version was sch80, too small a bore pressure tap/ftg mtd opp to branch, flush to bore barbs are CRUCIAL, use Eldon-James 1/2"NPT x 1/2" single barb, no mods connection tubing is 3"OAL, 3/8"ID pushed over the 1/2" barb (each side of DUT) all insulated, I'll take a pic later |
11-22-2005, 06:05 PM | #6 |
Responsible for 2%
of all the posts here. Join Date: May 2002
Location: Texas, U.S.A.
Posts: 8,302
|
Many thanks!
|
11-22-2005, 11:27 PM | #7 |
Cooling Savant
Join Date: Oct 2001
Location: Wigan UK
Posts: 929
|
Am I right in saying that,now, at 25c Environment Ch and 25c Coolant- Entry- T-joint there is a ~ 0.02drop in temp across the "6Tube +bits"'s DUT.
Previously (without? neoprene(inter-alia?) was ~0.06c drop Amongst previous data(link) making some sense of Series1 Which had taken to be 0.52"ID?(cf 0.93"ID? of 3/4"Ts) Crudely analysing in terms of Ev=(1-a)V^2/2gJCp and Ec=(Tw-Tx)/cosh[Rx*sqrt(h*Ap/kAp)] Edit: All crap.These are the errors which should cancel if using eqi-sized "in" and "out" T-things Have you abandoned smaller Ts? Thought had possibilities due to better mixing(Re ^) and if equalized to 3/8"ID of tubing,possible, less disturbance of flow. Apogee'ed brain needs time to re-acclimatize. Possibly no difference, possibly it is jumping out of the frying-pan into the fire. Probably best, for own sanity, to take hint from others and jump out of both.. Last edited by Les; 11-23-2005 at 01:57 AM. |
11-23-2005, 01:04 AM | #8 | |
Cooling Savant
Join Date: Sep 2003
Location: Vallentuna, Sweden
Posts: 410
|
Quote:
Bill I need to get this straight, having difficulty re-engaging brain. I don't really understand. Where is the water temperature measured? The correction would be for the RTD in the "T" piece? I guess I am wondering which is right... The hydraulic behaviour is I think real (that said without any deep investigation at all) and therefore requires no correction, just a statement of absolute water temp. ? I am not at all up to speed here. |
|
11-23-2005, 05:33 AM | #9 | |
Cooling Savant
Join Date: Oct 2001
Location: Wigan UK
Posts: 929
|
Quote:
This revealed apparent Dissipation by coolant at dT=0 and Absorbtion at dT=5.5 and dT=11.5 ? As shown in attachment? Excellent and I like* it if I have + and - signs correct. * Not sure,yet, about forcing me to use formulas |
|
11-23-2005, 08:21 AM | #10 |
CoolingWorks Tech Guy Formerly "Unregistered"
Join Date: Dec 2000
Location: Posts: 2,371.493,106
Posts: 4,440
|
Les
yes, but there may be an offset - I need to think for a bit on it may be ok ?? thinking of the swapped offset described in "C" Incoherent the instrumentation 'cross' is actually a modified tee, to function as a cross, with the branches having a RTD port and a pressure tap the crosses are on either side of the DUT and are shown bare in the first post and as now insulated below the measured difference is a steady state avg (peak to peak / 2) between the inlet and outlet (from the RTDs within the crosses, each corrected per its cal on that inst) equip kinda pushed to get these values want the spreadsheet ? EDIT the temps are NOT a simultaneous difference but independent max-mins, the lag varies with velocity lag Last edited by BillA; 11-23-2005 at 09:00 AM. |
11-24-2005, 10:17 AM | #11 |
CoolingWorks Tech Guy Formerly "Unregistered"
Join Date: Dec 2000
Location: Posts: 2,371.493,106
Posts: 4,440
|
well, there was an offset in the temp data; the coolant RTD corrections were reversed !
data is a lot more sensible now spreadsheets sent to those interested |
11-24-2005, 11:09 AM | #12 | |
Responsible for 2%
of all the posts here. Join Date: May 2002
Location: Texas, U.S.A.
Posts: 8,302
|
Quote:
|
|
11-24-2005, 11:29 AM | #13 |
Cooling Savant
Join Date: Oct 2001
Location: Wigan UK
Posts: 929
|
Just in time,about to look at 25c vs 15.6c.
Will still do,hopefully before Friday pub, but eases considerably. I note: considerable scatter at 25c(R^2=0.073), worse? at test coolant 30 or 35c, but doubt is relevant level back at no-neoprene level, consistent with my thoughts on still air vs closed cell neoprene Ta for spreadsheet |
11-24-2005, 11:56 AM | #14 |
CoolingWorks Tech Guy Formerly "Unregistered"
Join Date: Dec 2000
Location: Posts: 2,371.493,106
Posts: 4,440
|
could re-run but doubt sig chg
ok so far, but the air/liquid cal ? is unanswered (in my mind, all caled in flowing water for now) sent Ben |
11-24-2005, 12:08 PM | #15 | |
Cooling Savant
Join Date: Oct 2001
Location: Wigan UK
Posts: 929
|
Quote:
2)It is not - looking there but in terms of W 3) Also happy. |
|
11-24-2005, 01:19 PM | #16 |
Cooling Savant
Join Date: Oct 2001
Location: Wigan UK
Posts: 929
|
Think, you are now set up and calibrated to test the "6" of Tube" with any fan you wish.
Initial thoughts say you have to repeat the procedure with each radiator(with auxiliaries attached). Hope I am wrong |
11-24-2005, 02:21 PM | #17 |
CoolingWorks Tech Guy Formerly "Unregistered"
Join Date: Dec 2000
Location: Posts: 2,371.493,106
Posts: 4,440
|
yea, needs to be redone
agree that expressing in W is more understandable for most (but |
11-24-2005, 02:22 PM | #18 |
CoolingWorks Tech Guy Formerly "Unregistered"
Join Date: Dec 2000
Location: Posts: 2,371.493,106
Posts: 4,440
|
.
Last edited by BillA; 11-25-2005 at 08:28 AM. Reason: dup |
11-25-2005, 02:48 AM | #19 | |
Cooling Savant
Join Date: Oct 2001
Location: Wigan UK
Posts: 929
|
Quote:
Last edited by Les; 11-25-2005 at 06:10 AM. |
|